Analytical Measurement Uncertainty and Method Validation

Question 1:

In each question below, identify the most appropriate type of distribution that is being referred to. Then, calculate the standard uncertainty.

(a) Specification from the manufacturer for a 10 mL volumetric flask is 10 ± 0.025 mL. What is the standard uncertainty in the volume of the flask?

(b) A method specifies that a test portion of 1.000 ± 0.010 g is weighed for analysis. In subsequent calculations, the mass is treated as 1.000 g. What is the standard uncertainty attributable to this assumption?

(c) The calibration certificate for an analytical balance states that the measurement uncertainty is ± 0.000035 g @ 20 g load. The uncertainty is based on a confidence interval of 95%. What is the standard uncertainty of weighing?

(d) A dieldrin reference standard is described by the supplier as "99% min". Estimate the standard uncertainty of the purity.

(e) A calibration mass is certified as 0.05003 g ± 0.02 mg. What is the standard uncertainty of the mass in grams and in mg?

(f) The standard deviation of repeat weighings of a 50 mg mass is 0.033 mg. What is the standard uncertainty of a single weighing?

(g) Five absorbance readings were 0.123, 0.125, 0.128, 0.120 and 0.122.
(i) What is the standard uncertainty of a single absorbance reading?
(ii) What is the standard uncertainty of the mean absorbance?

Question 2:

(a) Calculate the combined standard uncertainty for the volume of iso-octane in a 100 mL volumetric flask from the following information.
- The results of 10 fill and weigh experiments on a 100 mL Class A volumetric flask had a standard deviation of 0.017 mL.
- Specification from the manufacturer for the flask is ± 0.1 mL.
- The coefficient of volume expansion for organic solvents, α, is 1x10-3 ºC-1.
- The mean laboratory temperature was 21.5ºC. The standard deviation of its variations was 2.3ºC.
NOTE: Temperature has an effect on the amount of the solvent contained in the flask, because of high coefficient of thermal expansion of organic liquids.

(b) A 5 mL pipette is used to dispense an organic solvent. Calculate the standard uncertainty in the volume of liquid delivered by the pipette.

- 10 repeat dispensings of liquid had a standard deviation of 0.005 mL.
- Tolerance stated by the manufacturer for the pipette is 5 ± 0.015 mL
- The coefficient of volume expansion for organic solvents, α, is 1x10-3ºC-1
- The mean laboratory temperature was 21.5ºC. The standard deviation was 2.3ºC.

(c) A method requires 50 mg of a dieldrin standard to be weighed out on an analytical balance. What is the standard uncertainty in the mass of the compound?
- 10 replicate weighings of a 50.03 ± 0.02 mg certified mass had a mean 50.039 mg and standard deviation of 0.033 mg.

(d) A primary stock solution is prepared by dissolving approx. 50 mg of dieldrin in iso- octane and making up to 100 mL in a volumetric flask.

I. Calculate the concentration of the solution in mg mL-1.

II. What is the standard uncertainty in the solution concentration?
- 50.3 mg of dieldrin was weighted out. The standard uncertainty due to mass as calculated in Question 2(c).
- The purity of the standard was quoted by the supplier as being 99% minimum. The standard uncertainty for this was calculated in Question 1(d).
- The standard uncertainty due to the 100 mL of iso-octane liquid was calculated in Question 2(a).

Question 3:

A standard solution of cadmium was prepared by dissolving cadmium metal in a nitric acid solution. The certificate of analysis for the cadmium metal stated that the purity was "0.9999±0.0001". The report also stated that, "the reported uncertainty is an expanded uncertainty with a coverage factor of 2, which gives a level of confidence of approximately 95%". In an analysis, 100.28 mg of the cadmium was weighed out, and the standard uncertainty on this mass was determined to be 0.05 mg, the associated degrees of freedom with this determination was 19. A calibrated 100 mL volumetric flask was used to prepare the solution to volume; the standard uncertainty on this flask was determined to be 0.07 mL. The associated degrees of freedom with this determination were 9. Using this information,

(a) Determine the concentration of cadmium in the standard solution.

(b) Determine the standard uncertainty of the concentration of this solution.

(c) Use the Welch-Satterthwaite equation to estimate the effective degrees of freedom associated with this standard uncertainty (note, you will need to estimate the degrees of freedom associated with the purity of the cadmium).

(d) Determine the coverage factor to be used to convert the standard uncertainty into an expanded uncertainty (assume a 95% confidence level) and report the result and its expanded uncertainty.

(e) The Eurachem guide on measurement uncertainty recommends using a coverage factor of 2 to calculate an expanded uncertainty with 95% level of confidence. Comment on this assumption compared to the coverage factor that you determined in (c) and (d).

Question 4:

In a laboratory, a standard solution must be prepared based on an aqueous solution of CH3COOH of concentration 4% (w/w) using the following procedure.

Forty millimetres of a specified stock solution of 100% (v/v) (standard deviation 0.5% (v/v)) is pipetted into a 1 L volumetric flask using a class A 20 mL pipette and the flask is filled with water. The difference between the laboratory temperature and the calibration temperature of the pipette and volumetric flask is not more than ±2°C.

The manufacturer's calibration data of the volumetric flask and the pipette and the standard deviation of the manual operations, obtained by earlier tests in the laboratory, are tabulated
below. The coefficient of volume expansion of water is 2.1x10-4°C-1.

Calibration data at 20°C

1 L volumetric flask ±4 mL
20 mL class A pipette ±0.03 mL

Standard deviation of the manual operations
1 L volumetric flask 1.5 mL
20 mL class A pipette 0.016 mL

Calculate the expanded uncertainty at the 95% confidence level. Include a cause and effect diagram in your analysis.

Question 5:

(a) To re-calibrate a 10 mL pipette, the volume of water (0.998207 g mL-1) was measured by 10 replicates giving the following results in grams.

9.985

9.982

9.981

9.982

9.972

9.984

9.980

9.985

9.983

9.981

What are the mean value and the standard uncertainty for a single pipetting step?

(b) The data set for the calibration of the photometric determination of nitrite is given below.

Standard

1

2

3

4

5

6

[nitrite] / mg L-1

0.05

0.10

0.15

0.20

0.25

0.30

Absorbance

0.1845

0.3197

0.4603

0.5895

0.7202

0.8501

The measured absorbance values for a sample are: 0.4892; 0.4886; 0.4895. Calculate the predicted average [nitrite] and the standard uncertainty of the sample.

(c) The specification for a 10 mL burette is quoted by the manufacturer as ±0.02 mL. Calculate the standard uncertainty under the conditions of a rectangular and a triangular distribution.

(d) According to the calibration certificate for a balance, the measurement uncertainty is
±0.0005 g at a 95% confidence level. Calculate the standard uncertainty.

(e) The standard deviation of repeated weighing of 0.1 g is calculated to be 0.00015 g. Calculate the standard uncertainty.

(f) The results obtained in a proficiency test of the determination of benzo[a]pyrene in drinking water according to a standard method are tabulated below. Calculate the standard uncertainty.

ctrue/mg L-1      caverage/mg L-1    RSD / %   sbetween-lab/mg L    swithin-lab/mg L-1

24                     19.05                  79.4                   4.92                    2.34

Chemistry, Academics

  • Category:- Chemistry
  • Reference No.:- M92844215
  • Price:- $60

Guranteed 36 Hours Delivery, In Price:- $60

Have any Question?


Related Questions in Chemistry

A hydrogen atom emits a photon as its electron changes from

A hydrogen atom emits a photon as its electron changes from n = 5 to n = 2. For the H atom, E n  = -2.18E-18*(1/n 2 ) J. What is the wavelength (in m) of the photon?  ( ) m The number of significant digits is set to 3; t ...

What wavelength of light nm is required to eject an

What wavelength of light (nm) is required to eject an electron with a speed of 2.0 x 10 6  m/s from a metal with a threshold energy of 3.4 x 10 -18  J?

A mixture containing 33 moles of no and 076 mole

A mixture containing 3.3 moles of NO and 0.76 mole ofCO 2  was allowed to react in a flask at a certaintemperature according to the equation. NO (g) + CO 2  (g) ↔ NO2 (g) + CO (g) At equilibrium 0.21 mole of CO 2  was pr ...

A dessert contains 13155nbspg of vanilla ice cream 90g of

A dessert contains 131.55 g of vanilla ice cream, 90g of fudge sauce, and 40.0 g of nuts. What is the total weight, in pounds, of the dessert?

A 250-ml sample of 120 m potassium chloride solution is

A 25.0-mL sample of 1.20 M potassium chloride solution is mixed with 15.0 mL of a 0.900 M lead(II) nitrate solution and this precipitation reaction occurs: 2KCl(aq) + Pb(NO 3 ) 2 (aq) -------> PbCl 2 (s) + 2KNO 3 (aq) Th ...

In the spring time a background ozone level of 290 ppbv was

In the spring time, a background ozone level of 29.0 ppbv was measured in Edmonton. Convert the mixing ratio of 29.0 ppbv ozone to units of molecules cm-3. Assume a pressure of 1.108 atm and a temperature of 12.5 ºC.

A chem major obtains a high-paying summer job and is given

A Chem major obtains a high-paying summer job and is given the task of determining the ethanol content (in vol %) of a commercial mouthwash sample. she decides to use gas chromatography because the experiment was so much ...

The density of air at stp is 129gl if your stoppered

The density of air at STP is 1.29g/L. If your stoppered the250mL flask contains 242mL of water(also the volume of air it would contain), calculate the mass of air in the flask under STP.

Calculate the dissociation constant of nh4ohaq if the

Calculate the dissociation constant of NH4OH(aq) if the degree of dissociation of 0.006 mol/kg solution is 0.053 and the activity coefficients of all species equal 1.

Aqueous lithium sulfate was mixed with aqueous strontium

Aqueous lithium sulfate was mixed with aqueous strontium chlorate, and a crystallized strontium sulfate product was formed. Consider the other product and its phase, and then write the balanced  molecular equation  for t ...

  • 4,153,160 Questions Asked
  • 13,132 Experts
  • 2,558,936 Questions Answered

Ask Experts for help!!

Looking for Assignment Help?

Start excelling in your Courses, Get help with Assignment

Write us your full requirement for evaluation and you will receive response within 20 minutes turnaround time.

Ask Now Help with Problems, Get a Best Answer

Why might a bank avoid the use of interest rate swaps even

Why might a bank avoid the use of interest rate swaps, even when the institution is exposed to significant interest rate

Describe the difference between zero coupon bonds and

Describe the difference between zero coupon bonds and coupon bonds. Under what conditions will a coupon bond sell at a p

Compute the present value of an annuity of 880 per year

Compute the present value of an annuity of $ 880 per year for 16 years, given a discount rate of 6 percent per annum. As

Compute the present value of an 1150 payment made in ten

Compute the present value of an $1,150 payment made in ten years when the discount rate is 12 percent. (Do not round int

Compute the present value of an annuity of 699 per year

Compute the present value of an annuity of $ 699 per year for 19 years, given a discount rate of 6 percent per annum. As